Orbitrap Tribrid Series 3.3 Patch
The unpatched version of software would try to set the pressure mode to “high” during the calibration on all instrument types. All non-Eclipse Tribrids will fail to reach this higher pressure mode (20 mTorr), and will return with an error. The patch fixes this defect, by setting the pressure mode to “standard” for this calibration. Please note that this defect is not present when using “FlexMix”. This is only an issue when using the Enfuvirtide calibration solution


Thermo Fisher Scientific