Improve determination of residual anions and organic acids for printed circuit board cleanliness

GaryHe
Team TFS
Team TFS

If you’re using ion chromatography to determine specific anionic contamination on the surface of printed circuit boards (PCBs), you want high resolution to ensure cleanliness.

 

Printed chip.jpgThermo Fisher Scientific teams recently demonstrated two techniques to improve resolution per the IPC-TM-650 Method. The team used a 2 mm i.d. Thermo Scientific™ Dionex™ IonPac™ AS11-HC column with 9 μm resin in two different techniques to modify the column selectivity to improve resolution.

 

Here’s a snapshot of the methods:

 

Method A

 

Columns

Dionex IonPac AG11-HC guard,

Dionex IonPac AG11-HC analytical, 2mm i.d.

KOH gradient, eluent pump

1 mM KOH (-5 to 9 min), 1-14 mM (9 to 16 min), 14 mM (16 to 23 min), 14-85 mM (23 to 30 min), 85 mM (30 to 32 min), 1 mM (32 min)

Eluent source

Dionex EGC-500 KOH eluent generator cartridge, Dionex CR-ATC 600 trap column, Dionex high pressure degasser module

Methanol gradient

None

Injection volume

5 μL

Flow rate

0.38 mL/min

Column temp.

20 ˚C

Detection temp.

20 ˚C

Detection

Suppressed conductivity, ADRS 600 suppressor, external water mode by pump 2 @ 0.38 mL/min, 57 mA

Conductance background

<0.5 μS/cm

System backpressure

~2,600 psi (100 psi = 689.5 kPa) with 800 psi backpressure tubing

Run time

38 min

 

Method B

 

Columns

Dionex IonPac AG11-HC guard,

Dionex IonPac AG11-HC analytical, 2mm i.d.

Proportioning pump 1

A: DI water for eluent generation
B: Methanol for methanol gradient

KOH gradient, pump 1

1 mM KOH (-5 to 17 min), 1-15 mM (17 to 24 min), 15 mM (24 to 35.3 min), 15-60 mM (35.3 to 54.6 min), 1 mM (54.61 -55 min)

Eluent source

Dionex EGC-500 KOH eluent generator cartridge, Dionex CR-ATC 600 trap column, Dionex high pressure degasser module

Methanol gradient, pump 1

2% (-5 to 24 min), 2-10% (24 to 29 min), 10-15% (29-51 min), 15-2% (51 to 54.6 min), 2% (54.6 to 55 min)

Injection volume

5 μL

Flow rate

0.38 mL/min

Column temp.

30 ˚C

Detection temp.

20 ˚C

Detection

Suppressed conductivity, ADRS 600 suppressor, external water mode (DI) by pump 2 @ 0.38 mL/min, 57 mA

Conductance background

~0.5 μS/cm

System backpressure

2,400-3,000 psi

Run time

60 min

 

These two methods are alternatives to the method presented in Application Note 1163: Determination of anions on the surface of printed circuit boards by IPC-TM-650 Method 2.3.28 using HPIC.  Compared to that method, both methods had reduced system pressure but some loss of resolution. The resolution of these two methods may be sufficient for the analytical goals of the PCB industry laboratory.

 

Learn more: To see the full method details and review the list of equipment, reagents, software and consumables, read this application brief.