Girón AJ, Deventer K, Roels K, Van Eenoo P.
Anal Chim Acta. 2012 Apr 6;721:137-46.
A new doping control screening method for the analysis of diuretics and stimulants using ultra high pressure liquid chromatography–high resolution Orbitrap mass spectrometry has been developed. The screening was performed in full scan MS with scan-to-scan polarity switching which allowed to detect more than 120 target analytes. Sample preparation was limited to 10-fold dilution of the urine into the internal standard solution followed by injection. Total run time per sample was 10 min. Validation of the method yielded detection limits for diuretics between 25 and 250 ng mL−1 and for stimulants between 5 and 500 ng mL−1. The screening method has been implemented in routine doping control.
http://www.sciencedirect.com/science/article/pii/S000326701200219XDepartment of Analytical Chemistry, University of Extremadura, Badajoz, Spain.