Institut de Physique du Globe de Paris, Sorbonne Paris Cité, Université Paris-Diderot, UMR CNRS 7154, 1 rue Jussieu, 75238, France
Journal of Analytical Atomic Spectrometry (2014), V29, pp1698-1707, doi: 10.1039/c4ja00098f
A fully automated direct injection nebulizer (d-DIHEN) for MC-ICP-MS isotope analysis: application to boron isotope ratio measurements
Pascale Louvat, Julien Moureau, Guillaume Paris, Julien Bouchez, Johanna Noireaux and Jérôme Gaillardet
This work presents a fully automated setup for using direct injection nebulization as an introduction system for solution measurements by MC-ICP-MS, here applied to boron isotopes in pure boric acid solutions and natural samples. In this setup, a direct injection nebulizer (d-DIHEN) is plugged into the plasma torch without any spray chamber, and an automated 6-port valve interfaces the nebulizer and the autosampler. The advantages of a d-DIHEN for boron isotope ratio measurements are high sensitivity and short washout times, allowing for sample–standard bracketing (SSB) measurements at a higher rate than spray chambers.
  • MC-ICPMS
  • Neptune
  • B isotopes
  • Direct Injection