Hi Everyone,

I have noticed the presence two mysterious peaks in almost every MS3 spectra acquired under RTS-SPS-MS3 mode on Eclipse.. These peaks are highlighted in the pic here.. I see them with both TMT and TMT Pro labels across different types of samples, and even with the Thermo Yeast TKO TMT Standard runs.. However, these peaks are not seen in TMT-MS2 acquisitions.. Have anyone else noticed this?? Any suggestion on where to start the investigation to uncover the mystery?? All leads are welcome.. Although those mysterious peaks are not impacting the outcome of the experiment, the mystery is kind of unsettling until it get resolved.. I am ready with my Sherlock hat on 😉

Thanks,

Bini

image002.jpg

2 REPLIES 2

Molly_I2
Community Manager
Community Manager

Hi @ItsBiniR - i will seek support from the team for you. Thank you so much for posting a question in our online community! We are grateful for this and will work to find you support - Molly

IofEyes1
Team TFS
Team TFS

These peaks are most likely noise, created by respectively, the ion trap RF (~150) and the quad RF (at ~180).  They would typically be eliminated from the displayed spectrum as exception peaks.  There is a diagnostic  “measure quad in Orbitrap” that would measure that frequency independently and update the exception list. Also exception peaks are typically hidden in QualBrowser/Freestyle, but can be viewed/hidden if this setting is changed.

mzCloud - A State of the Art Mass Spectral Library

Some resources only sound too good to be true.

MXCloud
mzCloud is a state of the art mass spectral database that assists analysts in identifying compounds.
It is a freely searchable collection of high resolution accurate mass spectra that can be accessed online free of charge.

Try it today!