I am trying to analyze methane sulfonate ( mesylate ion); however, I discovered the formate ion has the same retention time(exactly). I have tried to lower the KOH mM to 5 still no separation. I see with column selection guide the AS 11 HC uses 1 mM. I thought the AS 19 4um could separate better( this is what I am utilizing). The fluoride acetate is really good with the 10 mM starting. I may try to acidify the sample with h2So4 and utilize the huge PKA difference between formic and methanesulfonic acid. Start out with 1mM for the first 5 minutes. This is a hard separation. Can someone help! Thanks a million.
Solved! Go to Solution.
03-20-2023 04:50 AM
@IC-man , there are ways focus on IC method development. For early eluters such as fluoride, acetate, formate and methanesulfonate, it is better to use IonPac AS15 or AS18 although AS11-HC can also work. IonPac AS19 format columns are better for late eluters such as phosphate, perchlorate, iodate etc., not really for separation of early eluters.
Virtual column is a good starting point: https://appslab.thermofisher.com/VirtualColumnOnline
Other tips I have is that column temperature can sometimes change the retention times of your analyte.
Here are some method development tips: 5 tips to develop & improve your IC Method.
Watch out for the 2023 Global IC Symposium, coming soon!
I hope that helps you.👍
03-23-2023 07:54 PM
Thanks for the help!!!!!!!!!!! This resolved the issue I ordered a AS18.
03-19-2023 07:33 PM
Analytical instrumentation:
ICS6000
AS19 4um (4mm)
DRS600 suppressor
25uL loop
10mM 0-10min KOH
10-45mM 10-45min KOH
1.0mL/min
conductivity
03-20-2023 04:50 AM
@IC-man , there are ways focus on IC method development. For early eluters such as fluoride, acetate, formate and methanesulfonate, it is better to use IonPac AS15 or AS18 although AS11-HC can also work. IonPac AS19 format columns are better for late eluters such as phosphate, perchlorate, iodate etc., not really for separation of early eluters.
Virtual column is a good starting point: https://appslab.thermofisher.com/VirtualColumnOnline
Other tips I have is that column temperature can sometimes change the retention times of your analyte.
Here are some method development tips: 5 tips to develop & improve your IC Method.
Watch out for the 2023 Global IC Symposium, coming soon!
I hope that helps you.👍
03-23-2023 07:54 PM
Thanks for the help!!!!!!!!!!! This resolved the issue I ordered a AS18.